The mass flow meter shows a negative value

2018-05-31View Original

Thread Content

Please, experts, explain it to us
Reply #22018-05-31
If a check valve is not installed in the appropriate position during the installation of the mass flow meter, backflow can occur, which can lead to the phenomenon you are observing. Is there a check valve installed in front of your mass flow meter? Zero drift could also be the cause; close both valves to prevent the fluid from moving, and then perform a zero calibration
Reply #32018-06-01
First, check whether the pressure at the outlet of each gas cylinder is the same; when three flow meters are connected in series, if the pressure of the gas exiting the cylinders differs, the gas with higher pressure will push against the outlet of the flow meter with lower pressure, and in some cases this can even result in a negative value being displayed by the flow meter. It is recommended to install a one-way valve behind each flow meter.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.