I would like to ask the instrumentation engineers: Can a mass flow meter be used for measuring liquefied petroleum gas?
Thread Content
As the title suggests, mass flow meters generally do not provide accurate measurements in the presence of gas; in some cases, degasers need to be installed. I would like to ask instrument engineers: can mass flow meters be used for liquefied petroleum gas? How do you all consider this in the design process? Thank you!(1) Key parameters:
– Mass flow accuracy: ±0.002×flow rate ± zero drift
– Density measurement accuracy: ±0.003 g/cm3
– Density measurement range: 0.5–1.5 g/cm3
– Temperature measurement range: ±1°C
(2) Sensor-related data:
– Ambient temperature: –40–60°C
– Medium temperature: –50–200°C
– Explosion protection rating: iBⅡBT3
– Associated equipment: Compatible transmitter
(3) Transmitter-related data:
– Operating temperature: 0–60°C
– Relative humidity: Below 95%
– Power supply: 220±10% VAC, 50 Hz or 24±5% VDC, 40 W
2. Installation and use of the mass flow meter
Example 1: To support the expansion and renovation of our company’s gas distillation facility, we carried out the design and construction of the propylene loading system in early 1998. For measuring the propylene flow rate, mass flow meters produced by Taihang Instrument Factory were used. After the facility started operating, the flow meters functioned properly; however, there was a significant discrepancy between the values indicated by these meters and those recorded by the electronic scale. Cause analysis: (1) Through discussions with process engineers, it was ruled out that excessive dry gas content in the liquefied gas (5%) could be responsible for the inaccurate readings of the instruments. (2) A set of instruments with the same specifications was removed from the gas fractionation unit and installed for observation; the phenomenon remained unchanged, ruling out the possibility that the flow meter was damaged. (3) To this end, we sent people to inspect the Zhongyuan Oilfield Refining and Chemical Company; the flow meters used in its liquefied gas loading system are all mass flow meters. Since they were put into use over two years ago, they have been functioning properly with accurate readings. Its process flow and instrument installation are similar to those in our plant. The flow diagram shows that this plant places special emphasis on the sequence of operations: once loading begins, valve d on the loading line is opened first; once the readings on pressure gauges c and f become equal, then valve g on the gas return line is opened. In our plant, however, both valves d and g were opened right from the start. As a result, the pressure on the loading line became much higher than the pressure inside the tank truck, which made it easy for some of the remaining propylene in the tank truck to be pushed back into the propylene tank, thereby causing measurement errors. With improved operating methods, the system now functions properly, achieving accurate quantitative loading and thereby improving work efficiency. Example 2: In order to support the renovation of our company’s gas fractionation unit, we carried out the design and construction of a liquefied gas desulfurization unit at the beginning of 1998. The flow meter used in this unit was a mass flow meter produced by the Flow Meter Branch of Taihang Instrument Factory. After the unit started operating, the flow meter did not function properly; it frequently returned to zero in terms of instantaneous flow rate, and it also displayed an alarm signal indicating the presence of entrained gases. Cause analysis: (1) Through joint discussions among the instrument manufacturers and process engineers, it was ruled out that an excessive amount of dry gas in the liquefied gas (5%) could be responsible for the inaccurate readings of the instruments. (2) A set of instruments with the same specifications was removed from the gas fractionation unit and installed for observation; the phenomenon remained unchanged, ruling out the possibility that the flow meter was damaged. (3) It has also been observed that the zeroing of the instrument is related to the medium pressure. To this end, we sent people to the experimental plant of Luoyang Petrochemical Engineering Company for inspection; the flow meters used in the liquefied gas output systems of its two catalytic units have been functioning properly for over two years, providing accurate readings. The pressure of the medium measured by one set of instruments is 1.0 MPa, while the pressure of the medium measured by the other set of instruments is 2.0 MPa. In other aspects, it is similar to our factory. The pressure at the liquefied gas outlet of the liquefied gas desulfurization unit in our plant is 0.6~0.9 MPa. It has been observed that when the medium pressure falls below 0.75 MPa, the instrument reading returns to zero instantly. Based on this, it can be determined that 0.75 MPa is the saturated vapor pressure of the liquefied gas produced by our company (it is difficult to calculate this value due to incomplete data; it can only be determined based on chromatographic analysis results). At this pressure, the amount of gas evaporated from the liquefied gas exceeds 5% (M/M), which causes the instantaneous flow rate indicated by the mass flow meter to drop to zero. After discussion with the process engineers and operators, it was decided to install the sensor of the mass flow meter at the outlet of the liquefied gas feed pump in the desulfurization unit; the pressure of the fluid at this location is at least 1.3 MPa, and it remains stable. After the modification, the mass flow meter operates stably and with high accuracy, achieving the desired results.