How should faults be diagnosed when problems occur in electromagnetic flowmeters during use?

2020-03-26 View Original

Thread Content

  How to troubleshoot problems with electromagnetic flowmeters: 1. Check whether the operating voltage is correct based on the location where it is used; 2. Is the sensor grounding reliable or damaged? ; 3. Is the sensor filled with medium, and are there any bubbles in the medium? ; 4. Is the reading correct, and is the measured value within the measurement range? 5. Are the terminal connections loose? 6. Has the flow sensor been submerged in water?   How to check if there is a problem with an electromagnetic flowmeter: 1. The instrument shows no display ; 2. Check whether the power supply is connected ; 3. Check whether the power fuse is intact ; 4. Check whether the supply voltage meets the requirements ; 5. Check whether the contrast adjustment of the monitor can be adjusted and whether the adjustment is appropriate.   The causes of failures in electromagnetic flowmeters include the following aspects: environmental factors ; Fluid aspects ; Installation aspect. Environmental factors: These usually include interference from stray currents in pipelines, strong electromagnetic waves in the surrounding area, and the magnetic field generated by large motors. For interference caused by stray currents in pipelines, good separate grounding usually yields satisfactory results; however, in the presence of strong stray currents (such as those in electrolysis plants, where the peak value of the alternating voltage induced across the two electrodes can sometimes reach 1 V), additional measures are required, such as insulating the flow sensors from the pipelines. Space electromagnetic wave interference generally enters through signal cables, and is usually protected using single-layer or multi-layer shielding.   Fluid-related reasons: The presence of uniformly distributed tiny bubbles in the liquid being measured generally does not affect the normal operation of an electromagnetic flowmeter. However, as the bubbles grow larger, fluctuations occur in the instrument’s output signal. When the bubbles are large enough to cover the entire electrode surface, their passage over the electrodes can cause the electrode circuit to be momentarily interrupted, resulting in even greater fluctuations in the output signal. When an electromagnetic flowmeter driven by a low-frequency square wave is used to measure slurries with a high solid content, slurry noise is also generated, causing fluctuations in the output signal. When measuring a mixed medium, if flow sensing is performed before the mixture has been homogenized, it will also cause fluctuations in the output signal. An inappropriate selection of electrode materials and the medium to be measured can also affect accurate measurements due to chemical reactions or polarization effects. The electrode material should be selected appropriately based on the instrument or relevant manuals.   Reasons related to installation: Faults are usually caused by an incorrect installation location of the electromagnetic flowmeter sensor; a common example is installing the sensor at points in the pipeline where gases tend to accumulate ; Or installed on a vertical pipe from top to bottom, evacuation may occur ; Or there is no backpressure behind the sensor, and the fluid flows directly into the atmosphere, resulting in a non-full condition within the measurement tube.   Methods for troubleshooting electromagnetic flowmeters: resistance method ; Current method ; Voltage method ; Substitution method ; Analog signal method ; Waveform method. Resistance method: (1) On/off status of the flow meter power fuse and excitation fuse ; (2) Continuity of the signal cable and excitation cable for electromagnetic flowmeters ; (3) On/off status and resistance value of the excitation coil of the electromagnetic flowmeter ; (4) Insulation resistance of the excitation coil of the electromagnetic flowmeter to ground ; (5) Electrode symmetry measurement ; (6) Insulation resistance of the electrode to ground ; (7) Resistance value of the flow meter power transformer. Current method: Measure the output current and excitation current of the flow meter. Voltage method: Determine the operating power supply. Substitution method: Swap the converter and amplifier board of the flow meter. Analog signal method: Use an analog signal generator to provide a flow signal in order to test the flow sensor. Waveform method: Test the waveforms at key points based on an understanding of the circuitry. The inspection items for electromagnetic flow meters include: visual inspection, testing of converter characteristics, calibration of measurement values, measurement of voltages across various components, measurement of insulation resistance, and verification of the circuitry. During the inspection and adjustment of instruments, zero drift makes it very important to adjust the zero point (\"online zeroing\" requires the flow of the medium being measured to stop, which is not easy to achieve in practice).   Therefore, online inspections often omit the checks related to the operation of sensors, and only perform calibration of the converter in order to compare the results of the online inspection with historical data and determine whether the instrument should continue to be used, repaired, or replaced. As for the sensors, whether they need to be replaced is decided based on the degree of degradation in the insulation resistance of the measured excitation coils. Verify whether there are any abnormalities in the electromagnetic flowmeter through online inspection. For pipelines where the medium flow cannot be stopped, the flow sensor and the converter are inspected separately; the converter is tested using analog signal generators and other standard instruments to ensure high calibration accuracy (which depends on the accuracy of the analog signal generators). The sensor is checked by testing the resistance of the electrodes in contact with the fluid, as well as the insulation resistance and copper resistance of the excitation coil including its connection cables. Additionally, the excitation current output by the converter is examined, and indirect methods such as checking the magnetic field strength are used. For pipelines where the medium flow can be stopped, access can be obtained through an access hole pre-installed near the sensor, to inspect the condition of fouling/deposition on the electrodes and lining and to clean them.   Zero-point check of the entire electromagnetic flowmeter: The measurement tube of the flow sensor should be filled with liquid with no flow present. In many industrial settings, such conditions are not available, so the zero-point check and adjustment of the entire device are skipped; instead, a separate zero-point check and adjustment can be performed on the converter. Technically, this is only meaningful after the sensors have been inspected, and it is ensured that the insulation resistance of both the sensor excitation circuit and the signal circuit is normal (including the cables); otherwise, the entire device cannot operate properly. Usually, the individual zero point of the converter is negative, and the value is also very small ; If its absolute value is greater than 5% of the full scale, an inspection must be carried out first; adjustments should be made only after the cause has been identified. Under normal circumstances, the difference between the zero point of the entire electromagnetic flowmeter and the zero point of the converter alone is less than 1%. In many cases, zero-point differences greater than 5% are caused by the user performing incorrect zeroing operations when the pipeline valves are not properly closed.   Inspection of the connection cable for electromagnetic flowmeters: This inspection involves checking the insulation resistance of each core of the signal line and the excitation line, as well as verifying whether the shielding layers are properly grounded.   Check of the electromagnetic flowmeter converter: This check involves using general-purpose instruments as well as analog signal generators compatible with the flowmeter model to replace the sensor in order to provide a flow signal for zeroing and calibration. Calibration includes zero-point check and adjustment, setpoint check, excitation current measurement, current/frequency output check, etc. It should be noted that the inspection values need to be compared with the previous inspection values (or factory settings) to determine whether there are any changes, and whether such changes meet the original measurement requirements.   Electromagnetic flowmeter sensor inspection: Check the copper resistance of the excitation coil ; Check the insulation resistance of the excitation coil ; Check the electrode contact resistance ; Check the polarization voltage between the electrode and the liquid ; Check the insulation of the signal circuit and the insulation between the excitation circuit and the signal circuit. For more information, please visit the company’s official website at http://www.yb1518.com/. Please keep this link when reproducing the content! http://www.yb1518.com/UploadFiles/2012717174126829.jpg

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.

Quick Links