How do you conduct a helium leak test?

2008-11-25View Original

Thread Content

How do you conduct a helium leak test?
Reply #22008-11-26
1: Evacuate the chamber that is to undergo the helium leakage test; 2: Install equipment for monitoring helium leakage, ensuring that it can communicate with the interior of the chamber; 3: Cover the chamber from the outside using a sealing material when evacuation is required; 4: Introduce helium into the chamber and adjust the sensitivity of the leakage detection device; 5: Evacuate the chamber, stay nearby, and listen carefully; 6: If no alarm is detected by the monitoring device within 40 minutes, the test is successful. Note: The monitoring device must be calibrated before installation.
Reply #32008-11-26
1: Evacuate the chamber that is to undergo a helium leakage test; 2: Install the equipment for monitoring helium leakage, ensuring that it can communicate within the chamber; 3: Seal the chamber from the outside once evacuation is complete; 4: Introduce helium and adjust the sensitivity of the monitoring device; 5: Evacuate the chamber again and listen carefully; 6: If no alarm sounds occur within 40 minutes, the test is successful. Note: The monitoring device must be calibrated before installation. The parts marked in blue have been modified; the parts marked in red don’t seem to make sense in Chinese, and I think they are incorrect.
Reply #42008-11-26
It seems there are specific experimental protocols and equipment; it was only carried out on-site last month
Reply #52008-11-26
A helium leak detection machine is required, which comes in internal pressure and external pressure methods. Generally, for small containers, the internal pressure method is used: a certain pressure of helium is filled inside the device, and then an inspection is carried out from the outside using a suction gun.
Reply #62009-02-10
Helium detection is briefly described in Section 18.9.5 of the book \"Training Manual for Pressure Vessel Design Engineers\" written by Senior Li Shiyu.
Reply #72009-02-11
I haven’t actually dealt with it yet; I’m just coming in to take a look. I have indeed seen the helium testing for valves
Reply #82009-02-11
Simply put: 1. Evacuate the vacuum, 2. Replace with nitrogen, 3. Fill with a certain percentage of helium. 4. Hold pressure 5. Detect leaks with a helium leak detector (the sensitivity should be adjusted in advance) Last edited by nimoo on 2009-2-11 14:41]
Reply #92009-02-11
As a supplementary note: although helium leak detection is very sensitive, it is difficult to identify the exact location of the leak; one can only gradually narrow down the possible area where the leak might occur. It is much easier to operate compared to ammonia leak detection; the leak location can be easily identified, and it offers significant advantages in terms of cost.
Reply #102009-02-23
First, here is a direct translation of the original text: 1: Vacuum the Chamber which required Helium leakage test; The vacuum chamber must undergo a helium leak detection test. 2: The equipment for monitoring helium leakage shall be installed, and it must be able to communicate with the interior of the chamber; When installing the helium leak detector, its probes should be able to reach inside the vacuum chamber. 3: Cover the Chamber with a seal material from outside when the vaccum in the requirement is achieved; Once the desired level of vacuum has been reached, use sealing material from the outside to seal all areas of the vacuum chamber that might be prone to leaks. 4: Feed the Helium and adjust the leakage deadline of the monitor; Supply helium gas and adjust the minimum sensitivity of the leak detector. 5: Vacuum the chamber, stand nearby and listen; Check the level of vacuum and the integrity of the seals, while also listening for any noises. 6: If no alarm is issued by the monitor within 40 minutes, the test is successful. If no alarm occurs after 40 minutes, the test passes. Note: The monitor shall be calibrated before installation. Note: It is necessary to calibrate the leak detector before using it. In fact, everything is already explained quite clearly; I assume this is part of an installation or maintenance manual provided by a foreign party, likely related to the vacuum level testing requirements for turbine condensers. What confuses you now is the lack of a helium leak detector; otherwise, it would be quite straightforward. It can be purchased domestically or borrowed from the thermal testing laboratory; many alternatives are available, such as measuring the rate of vacuum drop or the maximum allowable vacuum level. The method remains the same even if it is not a condenser. The testing method is merely a means; what matters is identifying gaps and ensuring that the indicators meet the standards
Reply #112009-02-24
Helium leak detection must be the most sensitive method for detecting leaks. I know that helium leak detection is used to inspect the tubes and tube sheets of steam generators in nuclear power plants (which function as U-shaped tube heat exchangers). It’s a pity I wasn’t there at the time and didn’t see it.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.