Thread Content
We are using a GC9800 gas chromatograph; there are two fans at the back of the device, and I don’t know which components they are used to cool down, as the manual doesn’t mention it. One fan is behind the control panel, and one is behind the column box. The fan behind the control panel of one machine is broken. Given that there are few samples but many chromatograms, it’s not clear whether it’s possible to use the chromatograph without this fan Thank you!
The function of the two fans is as follows: the column box fan is used to maintain a uniform temperature within the column box; it blows the hot air generated by the resistance wires, dispersing it evenly throughout the column box, with the temperature being controlled using thermocouples or a temperature control circuit; The latter fan is used to dissipate some of the heat. There is a valve between the two; when the temperature is slightly high, the valve opens a little, allowing the excess heat to be released. As the temperature rises further, the valve opens more widely, and during periods of high temperature it may even close completely. If cooled rapidly, the valve opens fully. Effects of the two: if the former does not rotate, the distribution inside the column box will be uneven, with higher temperatures near the fan ; If the latter fails, heat dissipation will be impaired. Controlling the temperature is difficult; when the temperature is unstable, the impact on analysis is obvious. These can be shut down for inspection. Open from the back of the chassis. If there is a circuit fault, contact the manufacturer and electrical technicians for repair. First, you need to determine what is causing the fan not to rotate Power supply disconnected? Burnt resistor? Control circuit burned out? If you have the ability, take a look first. You can also consult the manufacturer.
One fan is located behind the control panel; it’s broken and can still be used, but it’s necessary to contact the manufacturer to obtain replacement parts. Another one is behind the column box. : This one is broken; it’s recommended to shut down the system for repairs – it’s a serious issue.