The magnetic flap level gauge is installed on the outside of a large storage tank. How long is a section usually?

2009-02-25View Original

Thread Content

The magnetic flap level gauge is installed on the outside of a large storage tank. If you need to use several wires one after another, how long is one section usually?
Reply #22009-02-25
Do not exceed 1700mm. If it exceeds 1700mm, you can use 2 units.
Reply #32009-02-25
It is best to be around 1.5 meters. If it exceeds 1.5 meters, use two sections.
Reply #42009-02-25
If it is for on-site indication of liquid level in a large storage tank, you must first look at the accuracy requirements you require, and select a magnetic flap level gauge with a length within the measurement range within the accuracy range. If it is relatively high, generally one section is limited to 5 meters. It is best to use two meters for measurement if it exceeds 5 meters (I have also seen manufacturers producing 6 meters, but I personally think 5 meters is the best). This post was last written by wwxxss_1 edited on 2009-3-5 19:26 ]
Reply #52009-02-25
We have installed two 6-meter magnetic flaps. When installing, I think it should not exceed two, otherwise it will be very troublesome to install. Moreover, when the flaps are messy, you need to use a magnet to smooth them from top to bottom.
Reply #62009-02-25
I have used Kopler's 5 meter and 7 section. If it is a 6 meter tank, it is best to use two 3 meter sections. If the magnetic replica is too long, it will easily deform after being used for a long time, and the reading will be inaccurate.
Reply #72009-02-26
Others upstairs have given the answer, 3 to 5 meters. It is also worth noting that there must be a certain overlap between the two paragraphs.
Reply #82009-02-26
One of the storage tanks is 11 meters, and the other is 12 meters. I originally wanted to use 5 meters and 6 meters, and use two sections. However, I consulted the manufacturer and said that 6 meters might be too long and would be inconvenient for loading and transportation, so I switched to three sections. The 11-meter tank uses 4 meters, 4 meters, and 3 meters.; Use three 4-meter tanks for a 12-meter tank. I would like to ask everyone, is it better to use the original 2 sections, or divide it into 3 sections?
Reply #92009-02-26
3 paragraphs is good. If it is too long, the measurement will be inaccurate after being used for a long time. It is best to install it in sections. Someone mentioned it above, you can refer to it!
Reply #102009-02-26
“What does “certain overlap” mean?
Reply #112009-03-05
If you can use 2 sections, it is better not to use 3 sections. Generally, the length should not exceed 6m. The more openings on the tank, the more leakage points.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.