2009-12-10View Original

Thread Content

The chemical processing unit is equipped with a pH meter for online analysis; the required value for this meter is not less than 12.5. The pH level of the liquid being measured (mainly softened water containing a small amount of AlCl3 particles) is around 13. The operating pressure is 12 barg, and the operating temperature is 60°C. Imported pH probes are currently in use at the site, and the following problems have arisen: probes that are new or calibrated with neutral water give accurate readings; however, as time goes by, their readings start to deviate from the pH of the liquid being measured, with the deviation being downward. For example, the offline measurement shows a value of 12.8, while the online reading is 12.3. The online probe was removed and calibrated using standard solutions at pH=5 and pH=8.5, with accurate readings from the probe. After reinstalling the calibrated probe, it was able to provide accurate readings for 2 days, after which the issue of low readings occurred. We are developing new types of online pH probes that can be used in such environments; we ask our colleagues at sea to provide information on the optimization of existing solutions as well as details regarding these new online pH probes. Thank you!
Reply #22009-12-11
I don’t know what kind of electrodes you use. Exposing a glass electrode to alkaline solutions for an extended period has an impact on its hydration layer; it generally needs to be calibrated once a day during continuous use.
Reply #32010-01-02
This is caused by a phenomenon known as \"alkaline deviation\" that occurs during pH measurement. It’s no problem. Using a pH meter to measure strong alkaline solutions is no longer a very appropriate method; if there are other analytical methods available, those should be used whenever possible.
Reply #42010-01-13
I agree with what was said on the second floor. Also, I wonder if the medium used for measurement might cause deposits to form on the probe.
Reply #52010-01-13
Industrial pH electrodes can be used, such as antimony electrodes, etc

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.