What is the cause of baseline drift and how to solve it?
Reply #22010-03-11
Possible reasons: The temperature of the detector is too high and needs to be lowered appropriately.
Reply #32010-03-11
2# * * ple But my detector temperature never changes
Reply #42010-04-03
1 It may be column loss or contamination. You can replace the substrate, or remove one or two circles at the column inlet. 2 Inlet port contamination, clean the inlet port. 3 Inlet port septum flow.
Reply #52010-04-06
This is explained in the instrument manual. You should read the manual carefully.
Reply #62010-04-06
If it is caused by external interference, you should find the reason yourself.
Reply #72010-04-07
Air leakage, carrier gas flow rate imbalance, inlet contamination, impure carrier gas, etc. can all cause baseline drift and should be investigated one by one.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.