The process reflection analyzer had no readings; during on-site handling, it was found that it was a hardware alarm caused by the flame going out. The alarm was cleared and re-ignition was attempted, but the flame would not ignite. What is the reason? Help needed
Reply #22010-07-22
The carrier gas at the site, the sampling port – everything is normal
Reply #32010-07-22
Check the air, gas, and carrier gas.
Reply #42010-07-22
FID’s? The temperature of the incubator?
Reply #52010-07-24
The problem is solved; the ratio of air to carrier gas was set too low
Reply #62010-07-24
Reply to 4# salmon.work: FID stands for Hydrogen Flame Ionization Detector; it seems you’re not in the field of analyzers either
Reply #72010-07-26
It’s a detector for flame ions! Didn’t he say the fire was out? ! I think it’s this one. Is there any problem? I’m not an analyst, but I have bought analytical instruments.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.