HCBBS Forum (English)
Submit Chemical Projects / Find Solutions
Amplify Your Requirements on a Broader Chemical Platform *Engineering · Technology · Equipment · Solutions*
Submit Request

How to use a dew point meter when the gas flow rate is insufficient? Everyone, please help. Thanks

2010-07-29View Original

Thread Content

As the title suggests: How to use a dew point meter when the gas flow rate is insufficient? I used a vacuum pump to draw air from the exhaust port in order to increase the flow rate, but the readings obtained were inaccurate. Pulling air out with the pump and then feeding it into the dew point meter also gave inaccurate results. I really can’t think of any other solution; since there’s no pressure enhancement device, I was wondering if there are any other simpler methods. Thank you!
Reply #22010-07-29
For measurements using a dew point meter, it is necessary that the pressure and gas flow rate remain stable; data obtained by using a pump to draw air are inaccurate. It is recommended that you install a buffer device behind the pump and give it a try.
Reply #32010-07-29
Reply to 2# Egret Shoal: What exactly is that buffering device? I would appreciate some guidance. I’ve only been working for a year and lack experience, so please help me out!
Reply #42010-07-29
If the gas pressure and flow rate are not sufficient to push the cylinder open, pull the cylinder out by hand; measurement can be carried out as long as there is gas flow.
Reply #52010-07-29
I wonder what exactly you consider to be a very small amount of gas flow? I agree with what was said above. I measured the dew point while the pressure was still in use; I haven’t measured it at low flow rates.
Reply #62010-07-29
Reply to 1# zhangw.007: The pump delivers pulsed flow, so it’s naturally impossible to obtain accurate measurements. There are two methods: one is to lengthen the instrument’s vent pipe; Second, install a needle valve or pressure stabilizing valve after the pump; this way, measurements can be taken even when the flow rate is low. The true measure of one’s on-site analysis ability lies in their adaptability.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.