HCBBS Forum (English)
Submit Chemical Projects / Find Solutions
Amplify Your Requirements on a Broader Chemical Platform *Engineering · Technology · Equipment · Solutions*
Submit Request

Issues with the weighing function of the automatic packaging machine during retesting

2010-10-29View Original

Thread Content

The re-inspection automatic packaging scale is the Korean fac5500 model, with a capacity of 30 kg and a minimum resolution of 10 grams. The setting is to reset automatically; each time it is calibrated using a 10kg weight, the error is around 20 to 30 grams. The base zero point varies between 1220 and 1780. The problem now is that calibration has to be done almost every day. Has anyone experienced this phenomenon while using it? Or where might the problem lie? I would like to ask everyone for advice, thank you
Reply #22010-10-31
By calibration, do you mean calibrating the zero point or the measurement range? Is it a drift in the zero point, or is the displayed range not matching the actual value? Is it because the sensor’s precision isn’t high enough, resulting in large temperature drift?
Reply #32010-11-04
The instructions state that the scale wasn’t adjusted properly, and there were issues with its fixation, which is what caused the values measured by the load cell to change; Of course, it’s also possible that the quality of your load cell is poor, resulting in creep
Reply #42010-11-04
The instructions state that the scale wasn’t adjusted properly, and there were issues with its fixation, which is what caused the values measured by the load cell to change; Of course, it’s also possible that the quality of your load cell is poor, resulting in creep
Reply #52010-11-04
Which manufacturer do you use?
Reply #62011-03-30
If you are looking for information on quantitative packaging machines for chemical products. Feel free to get in touch~~~~~~~~ Add QQ: 1183991459; Wei Gong: 13978821703. If you have any questions related to weighing in the future – such as automatic quantitative packaging scales, weighing and batching systems, or automatic weighing control systems – you can contact me directly. We will do our best to help you resolve the related issues.
Reply #72011-04-01
Thank you for those who are still paying attention to this issue. Now the problem has been identified; it’s an installation issue. One is the issue of the scale body being horizontal, and the other is friction between the axis and the scale body. It has returned to normal now.
Reply #82011-04-01
Our company also uses this product, and when we encounter such problems, it is often the cylinder that is at fault
Reply #92011-04-02
Reply to 7# weimoushan: Are there any application records related to the automatic packaging of caustic soda?
Reply #102011-04-10
Check the balance of the scale (whether the four corners are unbalanced); Check whether the sensor’s performance is poor (measure resistance and voltage) ; Is there a problem with the connection wire? ;

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.