Experience with calibrating E+H cable-guided wave radar: The empty tank shows around 20%, and this remains the case even after normal calibration! Select menu 09 to reset and calibrate again; the tank height is 6.5 m, LN=5.6 m, E=5.4 m, F=5.2 m. The display shows D+L=4.6 m+0.6 m. Choosing mapping=5.2 m still doesn’t work! It shows D+L=4.6m+0.6m; go to menu 03 and select \"Truncated\". LN=5.6m, the probe does not cover the entire area – OK!
Reply #22011-05-04
:victory: Great! Studying*. . .
Reply #32011-06-24
LN=5.6m, E=5.4m, F=5.2m; it shows that D+L=4.6m+0.6m, and mapping=5.2m is selected. Could you please explain what these letters represent? LN? E? F?D?L?
Reply #42011-08-23
Study* ................
Reply #52011-08-23
The tank height is 6.5 M; the cable length is 5.6 m, the empty tank height is 5.4 m, and the full tank height is 5.2 M. Based on these figures, the instrument should indicate a height of 5.2 M. If 20% is shown, it is possible to adjust the reading directly to the desired height while the tank is empty, within the suppression range. If suppression still isn’t possible, check whether the lower end of the cable is grounded. It is recommended to use the E+H calibration software, as it provides a very clear echo curve, which makes it easier to identify where the instrument malfunction lies!
Reply #62011-08-23
The tank height is 6.5 M; the cable length is 5.6 m, the empty tank height is 5.4 m, and the full tank height is 5.2 M. Based on these figures, the instrument should indicate a height of 5.2 M. If 20% is shown, it is possible to adjust the reading directly to the desired height while the tank is empty, within the suppression range. If suppression still isn’t possible, check whether the lower end of the cable is grounded. It is recommended to use the E+H calibration software, as it provides a very clear echo curve, which makes it easier to identify where the instrument malfunction lies!
Reply #72011-08-24
It’s still hard to understand when it hasn’t been used

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.