What are the reasons for sudden zero reading in vortex flowmeters?

2011-05-19View Original

Thread Content

I would like to ask the experts: what are the reasons that cause vortex flow meters to suddenly return to zero?
Reply #22011-05-19
It is possible that debris wrapped around the sensor prevents the vortex detector from detecting vortices temporarily. . . You can tap on the pipe to observe if there is any change in the indication. If the problem persists, consider the issue with the preamplifier. . .
Reply #32011-05-19
If the low-flow cut-off setting is set too high, the flow rate will suddenly drop to zero when it is around that value. Of course, the problem with the flow meter itself cannot be ruled out either
Reply #42011-05-20
Consider the low-flow resection problem; you can try it online again

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.