2011-07-27View Original

Thread Content

Our unit’s gas chromatograph has developed a problem: after hydrogen is used for ignition, the flame goes out automatically after a short while. Please help – anyone who has encountered a similar situation? Thank you first.
Reply #22011-07-28
There are two possibilities: either the temperature of the hydrogen flame detector is below 100 degrees, or the ratio of hydrogen to air is incorrect. Please measure again; it’s possible that there is a leak in that area
Reply #32011-07-30
At the critical moment, turn off the air supply, increase the hydrogen flow, then ignite it. Once it’s lit, gradually increase the air supply while reducing the hydrogen flow. Once, a colleague couldn’t get a flame to start; they tried using a lighter, but it just wouldn’t work. And you know what happened? They used the sample inlet as if it were a flame source, so of course it didn’t work.
Reply #42011-07-31
Our chromatograph has automatic ignition
Reply #52011-07-31
When igniting, the proportion of hydrogen should be slightly higher
Reply #62011-08-01
Also, check whether there is carbon buildup in the FID and clean it first.
Reply #72011-08-03
It could be due to excessive tail gas flow
Reply #82011-08-04
It may be caused by an excessive nitrogen flow rate or too low a hydrogen flow rate, or it could also be due to insufficient air volume.
Reply #92011-08-04
Reply to 6# E Road With You: The detector has been cleaned; it was fine for two or three days, but now it’s back to its old state.
Reply #102011-08-05
It is usually due to an excessive air flow rate
Reply #112011-08-07
It’s possible that the temperature of the detector is below 100 degrees; the water produced by the combustion of hydrogen does not evaporate and instead remains as water droplets inside the detector, which is why the situation you described occurs

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.