Gas Chromatography – Frequent APC pressure overpressure faults

2011-10-07View Original

Thread Content

This post was last edited by Cultivator on 2011-10-8 00:21. The hydrogen used in our gas chromatograph is supplied by a hydrogen generator; however, due to a problem with the generator, the alkaline solution from it ended up entering the chromatograph, causing the APC pressure to exceed its limit frequently. What should we do?
Reply #22011-10-18
After removing the contaminated section of the pipeline, wash it thoroughly several times with pure water; if possible, dry it (for metal pipelines), otherwise replace the air inside with high-purity gas after the washing is complete. If valves or other precision fittings are contaminated, they can also be removed, washed with water first, then cleaned with anhydrous ethanol and dried afterward.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.