2018-03-28View Original

Thread Content

The electrolyte used is a pre-made product that does not contain pyridine; it is used for detecting trace amounts of water in oils. The current electrolyte is no longer transparent and appears cloudy, seemingly due to a high oil content. Will this affect the measurement of water content? Is it necessary to replace it?
Reply #22018-03-29
Manufacturers provide pre-prepared electrolyte that can be purchased; it can be installed directly and used right away. The instruction manual includes steps for replacing the electrolyte, and the manufacturer who supplied the instrument should have provided a few bottles of electrolyte along with it – check there.
Reply #32018-03-29
In both the Coulometric method and the volumetric method, it is best for the solution in the system to be clear and transparent. If the solution becomes turbid or even stratified due to the testing of multiple samples, this can affect the measurement results; in such cases, it is necessary to replace the solution with a new one. Whether using the Coulometric method or the volumetric method, simply follow the instructions for replacing the electrolyte
Reply #42018-03-31
Okay, thank you(*?????*)????
Reply #52018-04-10
It’s a simple task to drain the electrolyte and replace it with fresh one.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.