I would like to ask the experts: what causes the absorbance value to be negative in the spectrum of a sulfur ratio analyzer?
Reply #22018-04-11
The absorption background of the sample is lower than the reference background absorbance value when using the originally calibrated instrument. It is sufficient to use actual sample gas to remove hydrogen sulfide and sulfur dioxide, and then use it as a background reference for calibrating the instrument; do not use nitrogen or other pure gases as the reference zero gas.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.