I need expert guidance regarding issues with ModScan32 reading data

2017-09-18View Original

Thread Content

When I use ModScan32 to read the cumulative values of the gauge, if those values have more than 7 significant digits, they cannot be read properly and are instead displayed in scientific notation. For example: if the gauge shows an accumulated value of 32508125, reading it with ModScan32 displays it as 3.25e+007 ; Another example: the gauge shows an accumulated value of 5678.1632, while reading it with ModScan32 gives 5678.0000. Is this because the data type of the instrument is single-precision floating point? Is there any way to make ModScan32 read properly? I would be very grateful if someone who knows a lot about this could give me some guidance.
Reply #22017-09-18
The data type displayed can be changed; for something like your cumulative value, it is usually a 32-bit floating-point number
Reply #32017-09-19
It is a 32-bit floating-point number; I switched to hexadecimal mode to read it, and then converted the hexadecimal value into a decimal floating-point number, and the result was correct. Is the problem with the instrument or with the reading software when such questions arise?
Reply #42017-09-19
This is probably a problem with your software; I used to use one that could display floating-point numbers directly
Reply #52017-09-19
I use the ModScan32 software for reading, and this software supports floating-point reading.
Reply #62017-09-19
If there is another scenario, namely when the numbers are too long, it is recommended that you change the unit; if the original value is in KG, convert it to T – that will also work
Reply #72017-09-21
I’ve tried this method – I changed KG to T – but it had no actual effect. For example, if the gauge displays 35288.789T, the value read by ModScan is also expressed in scientific notation: 3.52e+004. Similarly, when the gauge shows 352.88789T, ModScan defaults the digits after the decimal point to 0, showing a value of 352. So changing the unit has no practical effect. Another issue is that real-time reading isn’t possible. For example, if the value is 352, then every time 2 is added, a new value is read, such as 354, 356, and so on. If the value is a four-digit number, like 3528, then every time 16 is added, a new value is read, such as 3544, 3560. If the value is a five-digit number, like 35288, then every time 256 is added, a new value is read, such as 35544, 35800. That’s also a problem.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.