What is the reason why the overload protection trips as soon as the laboratory muffle furnace starts heating?

2020-03-06View Original

Thread Content

A multimeter test showed no short circuit between the heating wire and the casing, but the circuit breaker trips as soon as power is applied and the temperature starts to rise
Reply #22020-03-06
I think there are two reasons: one is tripping due to overload, and the other is tripping due to electric leakage. We can check whether the reset button on the circuit breaker has popped up; if it has, then the tripping is caused by electric leakage. If it hasn’t popped up, it means the tripping is due to overload. Of course, it’s necessary to ensure that the circuit breaker being used is accurate and functional.
Reply #32020-03-10
A muffle furnace is a heating device, and its temperature should not be high when measured with a multimeter. However, when a muffle furnace is in use, the high temperature causes a decrease in the insulation of the electrical components, resulting in electric leakage. If it is confirmed to be a leakage trip.
Reply #42020-03-10
Could you explain the principle behind how high temperatures lead to a decrease in insulation performance? Thank you
Reply #52020-03-11
At a visual inspection, the insulation of the heating wire is no longer effective; the high temperature has caused a ground fault and a shutdown
Reply #62020-03-11
First, use an insulation megger to check the insulation level; many multimeters cannot detect leaks in insulation
Reply #72020-03-11
Thank you. It’s indeed impossible to measure it with a multimeter

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.