Issues with the meter’s rated flow rate and maximum flow rate?

2016-06-13View Original

Thread Content

When choosing a mass flow meter, the meter has a rated flow rate and a maximum flow rate – what is the difference between these two flow rates?
Reply #22016-06-13
The rated flow rate is the flow rate that the manufacturer states the flow meter can achieve, at which flow rate the measurements taken by the flow meter are accurate. The maximum flow rate is the highest flow rate that can be achieved during use, and it is lower than the rated flow rate.
Reply #32016-06-13
Agree with what was said above; when choosing a flow meter, it is advisable for its rated flow to be 60%-80% of the maximum flow rate
Reply #42016-06-13
The friend on the 2nd floor gave an incorrect answer. The rated flow rate refers to the flow value of the fluid under specified operating conditions, while the maximum flow rate is the highest flow that the flow meter can measure under those same specified conditions. Generally, the rated flow rate is selected to be 50~70% of the maximum flow rate.
Reply #52016-06-13
The rated flow rate is the flow rate that the manufacturer states the flow meter can achieve, at which flow rate the measurements taken by the flow meter are accurate. The maximum flow rate is the highest flow rate that can be achieved during use, and it is lower than the rated flow rate.
Reply #62016-06-13
Just think of it as the rated voltage; if it’s exceeded, the circuit gets disconnected: lol
Reply #72016-06-14
The flow rate is measured most stably and accurately when it is at 60-70% of the maximum value.
Reply #82016-06-14
The reply above is already very detailed, so I won’t say anything more
Reply #92016-06-22
May I ask which manufacturer’s flow meter it is? It seems that Emerson’s flowmeters don’t have a concept of rated flow rate. Additionally, a brief note on flow rate: the maximum flow rate indicated on typical mass flow meters corresponds to a flow velocity of around 20 m/s. However, the calibration devices used by various companies that comply with **calibration standards can achieve at most 11–12 m/s; even those from Emerson cannot reach this level. Therefore, the maximum range indicated on mass flow meters is, in the vast majority of cases, determined through theoretical calculations or by referring to values from others. Of course, this does not affect performance in most situations, as the actual flow velocities under operational conditions are far below 11 m/s. But if there is indeed a high flow rate, it is still recommended to have the manufacturer perform a real-flow calibration. If it is a water-related device, the manufacturer has the capability to do it.
Reply #102016-06-22
Generally, for maximum flow meters, a value between 1/2 and 2/3 of the rated flow is the most reliable choice

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.