Is the accuracy class of 0.15 for LNG mass flow meters referring to the indication error?

2023-05-26View Original

Thread Content

I see that according to 7.1.1.3 of JJG1114-2015, the accuracy level of the calibration device is required to be no less than 0.15 grade, while many specifications related to calibration devices require that the indicated error of such devices be no greater than ±0.5%. Are these two requirements contradictory? Does the accuracy level of a calibration device refer to its indicated error?
Reply #22023-05-26
The accuracy class of 0.15 for LNG mass flow meters refers to the precision level of the instrument itself, rather than the indication error. This level of accuracy encompasses all error factors, such as the instrument’s linearity, repeatability, etc. The indication error represents the difference between the value indicated by the instrument and the true value; it is usually expressed as a percentage. Therefore, the requirements for calibration devices specified in JJG1114-2015 are not contradictory to the requirement in other standards that the indication error shall not exceed ±0.5%. The accuracy class of a testing device is a comprehensive indicator that takes into account the influence of various error factors, whereas the indication error is the difference between the value output by the instrument and the true value. .

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.