The hydrothermal reactor broke down after the experiment. Seeking advice

2017-04-17View Original

Thread Content

An appropriate amount of selenium powder (Se) was added to the hydrazine hydrate solution, and after stirring for about 1.5 hours, the mixture was transferred to a hydrothermal reactor. It was then heated at 160 degrees Celsius for 6 hours. Upon completion of heating, it was found that the reactor had been damaged and could no longer be opened. What’s going on here
Reply #22017-04-18
Hydrazine hydrate decomposes into a large amount of gas at high temperatures, and the high pressure causes the lining to deform; my 50 ml reactor, together with 15 ml of hydrazine hydrate, was damaged as a result. You can try reducing the volume of hydrazine hydrate used

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.