HCBBS Forum (English)
Submit Chemical Projects / Find Solutions
Amplify Your Requirements on a Broader Chemical Platform *Engineering · Technology · Equipment · Solutions*
Submit Request

Vortex flowmeter fluctuations

2017-03-18View Original

Thread Content

This post was last edited by Golden Heart on 2017-3-19 at 13:35. This flow meter is not showing any flow rate; the value stays within the range of 200 kg/h to 5000 kg/h! After removing the watch, it shows zero! Then there are fluctuations again after installation! It’s an empty pipeline now! Help! ! The problem isn’t solved yet! ! !
Reply #22017-03-18
Does the pipeline vibrate? Vibration can cause fluctuations in flow rate!
Reply #32017-03-18
The vibration of pipe 3# is not very severe!
Reply #42017-03-19
We have also encountered this situation, where we simply replaced the board
Reply #52017-03-19
Is the grounding of the vortex flow meter proper? Are there any electromagnetic interferences in the vicinity? ! !
Reply #62017-03-19
If there is vibration in the pipeline, small-signal rejection can be implemented without affecting its operation. There is an option for seismic resistance in the parameter settings, which can increase the seismic resistance coefficient. You can contact me for details; we also produce vortex flowmeters.
Reply #72017-03-19
The pipeline may vibrate; when actual fluid is passed through it, it should function normally at high flow rates, and it can be left to flow on its own when not in use.
Reply #82017-03-20
Try reducing the magnification factor or sensitivity of the amplification board~~ Also, make sure there is no strong electrical interference or high-frequency noise in the surrounding area, and that the vibration of the pipes complies with the installation specifications~~
Reply #92017-03-20
The sensitivity is too high; consult the manufacturer – lowering it should solve the issue.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.