A horizontally installed mass flow meter with an oil drain pipe at the back showed a flow rate that was lower than the actual value during operation. The flow meter displays the flow rate when it is unloaded. Is there any way to improve it?
Reply #22017-10-29
Check if there is a leak or vaporization of the medium. Whether there is internal leakage in the valve, back pressure at the flow meter outlet
Reply #32017-10-29
Check if there is a leak or vaporization of the medium. Whether there is internal leakage in the valve, back pressure at the flow meter outlet
Reply #42017-10-29
Check if there is a leak or vaporization of the medium. Whether there is internal leakage in the valve, back pressure at the flow meter outlet
Reply #52017-10-29
If the medium is viscous, clean the flow meter and check for vibrations and installation issues
Reply #62017-10-29
An under-filled tube is a major concern; when the tube is not fully filled, the density of the mass flow meter changes erratically, resulting in inaccurate measurements. The improvement method involves raising the oil supply pipe at the rear above the level of the mass flow meter; if possible, make a U-shaped bend in the pipeline at the rear.
Reply #72017-10-29
Saturator reset to zero! Or to check whether the installation is level and whether there are any misaligned pipes!
Reply #82017-10-30
Check whether the installation method is correct. Here, the medium used for gas measurement contains very little water; it is installed horizontally with the measurement tube (sensor) positioned at the bottom. Flow is indicated when there is an excess amount of fluid, so we just need to install the measurement tube at the top
Reply #92017-10-30
Close the valves before and after the flow meter, and check whether the reading is zero. If not, it is experiencing zero drift, and the zero point needs to be recalibrated on-site. If the flow rate fluctuates greatly, as mentioned above, the pipes downstream may not be filled properly due to suction, and adjustments need to be made.
Reply #102017-11-01
Is it full? Adjusting certain buttons when there is material present can cause data migration
Reply #112017-11-01
For the side gas, switch to vertical installation

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.