Some time ago, the heating wire in the HACH CODMA*I digestion cell in our company broke. Due to the lack of spare parts, the instrument was out of service for a while (without power being cut off). Yesterday, a new digestion cell was installed; before the replacement, the CODMA*I device showed an I2C error. What kind of error is that? How to handle it. Thank you!
Reply #22018-10-15
As shown in the image, the maintenance program cannot be executed
Reply #32018-10-15
:Oh, there’s even such a mistake? Don’t you have the machine’s manual? If I remember correctly, the meaning of this code seems to be “internal bus error”; action to take: contact the customer’s maintenance department
Reply #42018-10-16
The user manual does not list this error code. In previous versions of CODMAX, if there was a bus error, it would simply indicate a bus error, and it was also possible to define each component individually. In this current version, an I2C error is reported, and the maintenance routine is not even triggered; that’s why I have no other choice but to turn to all of you for help.
Reply #52018-10-16
Using the code “I2C” to search for information on industrial buses, I2C is one of such buses; therefore, based on the relevant sections in the user manual, the possibility of an “internal bus error” is very high
Reply #62018-11-01
The I2C error has been resolved, thank you all. The reason was that the 12VDC voltage from the COD power supply was not being supplied, while the +/-5VDC outputs were working normally; since the monitor displayed normal signals at that time, no issues with the power supply were considered. :lol:lol:lol
Reply #72018-11-02
I found it myself, but it was by accident. I accidentally opened the display cover to check the power supply and found that the 12VDC power indicator light was off.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.