How to calibrate the zero point of a differential pressure flow meter?

2016-01-06View Original

Thread Content

Should the three-valve block be turned off or on?
Reply #22016-01-06
All three valves should be either open or all closed, and then the built-in exhaust device of the instrument should be turned on
Reply #32016-01-07
Close the two pressure tapping valves of the three-valve assembly, open the balance valve, and zero it out.
Reply #42016-01-07
Typically, differential pressure flowmeters are equipped with a low signal cut-off point; this point must be disabled first before performing the zeroing operation.
Reply #52016-01-07
First, close the valve on the positive pressure side and open the balance valve; then close the valve on the negative pressure side to zero the setting; After completion, open the valve on the positive pressure side, close the balance valve, then open the valve on the negative pressure side for normal operation.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.