How to determine the quality of the diaphragm box in a double-flange transmitter?

2016-10-12View Original

Thread Content

In the factory, there is a gauge diaphragm box that is dented, but it still has elasticity and the silicone oil is still present. If the load is removed, the gauge also returns to zero. Please, those who understand can explain – can this watch still be used?
Reply #22016-10-13
If it is determined that there is no leakage of silicone oil, as long as the diaphragm still has elasticity, it can be used weiterhin.
Reply #32016-10-13
If there is still silicone oil present, it shouldn’t leak; otherwise, it would seep out little by little. It’s recommended to check first (as someone else mentioned above)! Pressure vessels cannot afford the slightest carelessness: lol
Reply #42016-10-13
Check a few points to see the linearity and such; zeroing is just part of it

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.