Factors affecting paramagnetic oxygen

2017-01-21View Original

Thread Content

I’m using a paramagnetic oxygen analyzer; it suffers from significant issues with zero point and range errors. Even after calibration today, the errors remain large tomorrow. The instrument used is the Sifumei 5410A, with an oxygen measurement range of 98–100
Reply #22017-01-23
The paramagnetic oxygen analysis method is affected by multiple factors depending on the design structure of the analyzer. 1. Is the instrument installed horizontally with no mechanical vibrations? 2. Is the sample gas flow constant? 3. Are there any interfering components in the sample, especially nitrogen oxides? 4. Is the calibration process reliable? 5. Are there significant changes in environmental and sample temperatures?

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.