Does anyone know the steps for calibrating the Hash SC200?
This post was last edited by qugd on 2018-8-7 at 19:44. SC 200 is a smart controller from Hash Company, capable of connecting to all of Hash’s digital probes as well as some analog probes. Regarding the calibration method for the connected probe, it is specified in the corresponding probe operation manual, not in the SC200 manual; you need to consult the relevant probe manual. The calibration function is located in the probe settings menu; different probes have different calibration methods, and you can follow the steps indicated in the corresponding manual.
You can take a look at other products; they offer great value for money
I have found the manual for the corresponding probe, which contains the calibration steps; thank you for your guidance. Do you know anything about ORP tables? At the last Hash conference I attended, a technical expert said that calibrating ORP is pointless; it’s just a qualitative indicator. I am used to measuring the reverse osmosis feed water.
This post was last edited by qugd on 2018-8-17 at 10:06. It was him fartsing. ORP calibration is certainly meaningful; without calibration, how can the potentials of different reference electrodes be used to compensate for the actual potential, and how can small drifts in the reference electrode potential be compensated for? What’s that technician’s name? I can help you ask him.