I would like to ask fellow sailors: when calibrating instruments, by how much higher must the accuracy of the standard gauge be than that of the gauge being tested? Which standard has relevant provisions?
Reply #22018-11-16
According to the calibration specifications for the relevant instruments, there are requirements regarding the accuracy of the standard meters.
Reply #32018-11-17
The table on measurement transmission systems focuses specifically on the topic of measurement traceability
Reply #42018-11-17
The relative error of the standard table is one-third of the relative error of the tested table.
Reply #52018-11-19
The relative error of the standard table is 1/3 to 1/4 of the relative error of the table under verification; of course, this is also generally derived from the uncertainty.
Reply #62018-11-21
For pressure transmitters, the uncertainty of the calibration equipment should be 1/4 of the absolute value of the maximum allowable error of the pressure transmitter being tested. For pressure transmitters rated at 0.1 or 0.05 precision, this uncertainty should not exceed 1/3 of the absolute value of the maximum allowable error of the transmitter in question~

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.