For the Terrell TA-5000 analyzer, the zero-point gas is high-purity nitrogen, while the range gas has a concentration of 5 ppm. How should the zero point and range be calibrated?
Reply #22020-02-16
Calibrate it in the usual way – can’t calibrate it?
Reply #32020-02-16
The specific calibration method can only be found in the instructions; for the detection of trace amounts of CO2 at the molecular sieve outlet, it is not recommended to calibrate the zero point – only the measurement range should be calibrated

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.