Where is the stability of the PH sensor? What are the components that tend to malfunction most often when people use them on a regular basis?

2020-03-24View Original

Thread Content

I’ve seen many domestic manufacturers producing glass bulb pH electrodes, as well as those from E+H. The analog glass electrodes all seem identical to me. In actual use, is it the glass electrode itself that affects stability, or is it the transmitter circuit that has a greater impact on performance? Is it necessary to choose imported pH electrodes?
Reply #22020-03-24
In our on-site use, glass electrodes have a significant impact. The replacement frequency is relatively high. Brand: Mettolidoid
Reply #32020-03-25
How long does it generally take to calibrate a pH meter on-site? The time required
Reply #42020-03-25
pH meter electrodes are consumable items; the further the pH value is from 7, the shorter the lifespan of the electrode. The pH meter can have its probe calibrated at home, and it can be replaced directly on-site.
Reply #52020-03-26
For pH electrodes, aside from rupture, the biggest issues are those related to the liquid junction and the reference electrode. The liquid junction is blocked and scaled by medium contaminants ; Excessively harsh media cause contaminants to penetrate the reference electrolyte inside the reference electrode, and the damage or contamination of the silver wire can both lead to measurement drift. In most cases, the choice of electrode depends on the medium, with different reference systems being selected accordingly; of course, high temperatures, high pressures, or particle friction are also taken into consideration at times
Reply #62020-03-27
The verification electrode is fast to use, and its operation is very simple

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.