Process Engineering Division – Instrumentation and Automation Section – Daily Question – Question from 2020-06-21

2020-06-21View Original

Thread Content

Process Engineering Section – Instrumentation and Automation Section – Daily Question – Question from 2020-06-21: Multiple-choice question: 155. For Micro Motion mass flowmeters, when the actual pressure used is much higher than the calibration pressure, what should be the relationship between the measured mass flow rate and the actual flow rate? (Note: There is no pressure transmitter compensation at the site, and the flow coefficient has not been adjusted.) A、Less than the actual flow rate B、More than the actual flow rate C、The same as the actual flow rate D、It may be more than the original flow rate or less than it. Answer: A、Less than the actual flow rate

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.