The electromagnetic flowmeter does not display the flow rate

2016-03-27View Original

Thread Content

I recently purchased an electromagnetic flow meter for production use, with a range of 0–20 cubic meters per hour. I now need to measure a flow rate of 70 liters per hour, but the instrument does not display this value. Question 1: Is it because the traffic volume is too low to detect it? It’s still related to the dripping rate issue. 2. Can flow meters of the electromagnetic flow meter type really not measure such low values? Is it necessary to switch to a mass flow meter?
Reply #22016-03-27
How is the conductivity of the medium? Could you please provide the electromagnetic caliber? In electromagnetic flowmeters, there is a parameter called lower limit cut-off, which can be adjusted to a lower value or turned off
Reply #32016-03-28
Could you please provide the electromagnetic caliber? The electromagnetic flowmeter is estimated to have a minimum flow rate of 0–20 l/min for DN32, and measurement is required with the pipe completely filled. It is recommended to use another flow meter. For example: twin-screw flow meters, gear flow meters, rotor flow meters, etc.; mass flow meters are used when high precision is required
Reply #42016-03-28
First, look at this range: 0–20 cubic meters per hour. The required value now is 70 liters per hour, which equals 0.07 cubic meters per hour. It’s 0.3% of the full range, so it has already been cut off; the corresponding current is 4.048 mA. Do you think the DCS will be able to read this value...? Suggested solutions: 1. The range ratio of ordinary electromagnetic flowmeters is 1:100; setting the range to 0–0.2 cubic meters per hour may allow for accurate measurements (ensuring the pipe is full). 2. Choose an electromagnetic flowmeter with a small diameter and install reducers before and after the pipe. Small diameter ≈ small range. 3. This is a matter of selection; it has nothing to do with the type of flow meter at all. Choosing a mass flow meter with a range of 0–20 cubic meters per hour will not allow you to obtain any measurement data… 4. Apart from the methods mentioned above, there are no other cost-free solutions……
Reply #52016-03-30
First, ignoring conductivity, your flow rate is too low, so this electromagnet is not suitable. Secondly, you can choose a mass flow meter for this; don’t select one based on 20 m3/h – instead, choose one corresponding to your current flow rate of 70 L/H.
Reply #62016-03-30
70L/H? It doesn’t matter if they’re not satisfied; stop testing it – just give up now.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.