I’m seeking help from experienced forum users: what is the maximum daily pressure increase in a liquid nitrogen storage tank that is considered normal?

2017-12-08View Original

Thread Content

Our factory has just put into use a 50-liter liquid nitrogen tank, which contains around 40 liters of liquid nitrogen. However, the pressure seems to increase quite rapidly – by about 4 kilograms per day. Is such a rate of loss considered normal?
Reply #22017-12-09
The liquid nitrogen tank is equipped with a pressure-boosting device, in case the pressure-boosting system is turned off. Gaining 4 kilograms in a day is definitely not normal. If it keeps rising every day, causing the safety valve to activate, that is definitely abnormal. If the pressure increases slowly once it reaches a certain value, it indicates that the storage tank is functioning properly. It’s hard to make a judgment without knowing the specific circumstances.
Reply #32017-12-10
The automatic pressure increase system is not in use; visually, there is no frost on the pipelines related to this system, so internal valve leaks can be ruled out. Nevertheless, the pressure still rises by at least 2 kilograms per day, which suggests that the insulation of the tank is not adequate.
Reply #42017-12-10
The automatic pressure increase system is not in use; visually, there is no frost on the pipelines related to this system, so internal valve leaks can be ruled out. Nevertheless, the pressure still rises by at least 2 kilograms per day, which suggests that the insulation of the tank is not adequate.
Reply #52017-12-11
The pressurization system is turned off. Increasing the pressure by two kilograms per day must be a problem. 1. Vacuum issue in the storage tank, measured using a vacuum gauge. 2. The valve has a very slight internal leakage, which is difficult to eliminate. 3. The design of the interlayer pipeline is unreasonable. It’s best to let the manufacturer handle it. There’s a problem with the new tank. Could you tell me which manufacturer the storage tank is from?

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.