What type of instrument is suitable for a glycerin level switch?

2017-09-24View Original

Thread Content

For the renovation of the on-site equipment, there is a glycerin storage tank; foam is used in this tank, and the operating temperature is below 300 degrees. Previously, the liquid level was monitored manually, but with the technological upgrades, instruments are now needed to control the liquid level. I need help determining which type of level switch is suitable for detecting foam in glycerin levels; thank you!
Reply #22017-09-24
Using a differential pressure gauge or an internally inserted floating ball would be better
Reply #32017-09-24
Can a float level switch withstand a temperature of 300?
Reply #42017-09-24
This can be asked to the instrument manufacturer; with the current level of manufacturing and materials, there should be no problem
Reply #52017-09-25
For a double-flange differential pressure level gauge, if it is intended to control only the high and low level points, a fork switch can be installed at each of these levels.
Reply #62017-09-25
Glycerin produces foam – does it have an effect on the tuning fork?
Reply #72017-09-25
The tuning fork available now should allow for adjustment of its sensitivity. There should be a difference in the vibration frequency of the tuning fork when it is placed in foam versus glycerin; it’s sufficient to adjust it so that it responds when in glycerin but not when in foam
Reply #82017-09-25
For high-temperature float level switches, please refer to the following
Reply #92021-08-26
I have a glycerin storage tank at room temperature; can a magnetic flip-type level gauge be used for it?

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.