HCBBS Forum (English)
Submit Chemical Projects / Find Solutions
Amplify Your Requirements on a Broader Chemical Platform *Engineering · Technology · Equipment · Solutions*
Submit Request

E+H Promass300 series mass flow meters

2018-07-19View Original

Thread Content

Which companies in the offshore oil industry are using or have used 300-series mass flow meters? Are there any relevant operation guidelines available? Where can the analog output of this type of flow meter be set?
Reply #22018-07-19
You can contact the relevant personnel at E+H; they can send you an electronic version of the manual. Since the product comes with after-sales support, just ask them for help.
Reply #32018-07-19
I just started working at this company, and it’s other colleagues who are currently in charge of handling external communications. If I were to ask for the contact information of the service providers on my own, it could lead to misunderstandings; I checked online for flowmeters in this series, and it seems they aren’t widely used; there’s quite little information available online
Reply #42018-07-20
There is an operation manual on the E+H official website, and the debugging section of it contains detailed setup instructions. I looked around, but there’s no 300; however, there is an I 300. Although the models are different, these settings should not differ much. The attachment is the operation manual for Proline Promass I 300
Reply #52018-07-20
You can go to the diagnosis menu, find the simulation options, and simulate the output current.
Reply #62018-07-20
I went and tried it today; it’s the last item in the diagnostics – I hadn’t noticed it before; I was rather careless
Reply #72018-07-20
Why is 300 used so little? I have 80 – what’s the difference between 80 and that?
Reply #82018-07-20
Friend, what software did you use to take the screenshot?
Reply #92018-07-23
Promass 300 is the model of the new transmitter, while 80/83 are the older models; the 300 can be considered an upgraded version that replaces the 80/83. While 80/83 will be phased out gradually
Reply #102018-07-23
E+H’s own simulation software. Now, the 300 transmitters can be connected to an Ethernet cable, making it very convenient to adjust them through a computer web page.
Reply #112019-11-27
To connect to lines 25 and 26, set it to analog output in the signal settings.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.