I work in the reverse osmosis industry, where the conductivity of the water produced by reverse osmosis is around 100 us/cm. What type of flow meter would be suitable for use in such cases? Currently, an E+H electromagnetic flowmeter is being used, but according to the information I found, electromagnetic flowmeters are not suitable for low conductivity?
Reply #22018-08-16
A conductivity of 100 should be fine to measure using electromagnetic methods; if there are concerns, then use vortex flow meters
Reply #32018-08-17
For reverse osmosis produced water, HP51530 series plug-in rotameter flow meters are generally used. Flow sensor: HP51530-P0 (suitable for diameters 0.5–4", DN15–DN100) ; HP51530-P1 (suitable for 5-8" diameter, DN125-DN200) ; HP51530-P2 (suitable for diameters 10-36", DN250-DN900) ; Flow transmitter: H3-9900-1P(H3-8550-1P)
Reply #42018-08-17
There are indeed quite a few of those rotary flow meters; we use GF’s ones
Reply #52018-08-18
For electromagnetic flowmeters, the required conductivity is generally above 20 us/cm; values as low as 5 us/cm are also acceptable, so it’s safe to use a value of 100.
Reply #62018-08-19
Feel free to use this electromagnet; values of 5us or more work fine, as I have conducted tests. Don’t listen to those who suggest using GF rotors; it’s all just numbers that are generated, and upon verification, they’re all incorrect
Reply #72018-08-19
The electromagnetic flowmeters from Cologne can be used; in our field conditions, even when the conductivity fluctuates around 30 us/cm, measurements can still be taken properly.
Reply #82018-08-20
:) Electromagnetic flowmeters are more than capable of handling this task, and their accuracy is guaranteed; a conductivity level of over 5 us/cm is generally sufficient, while split-type flowmeters may require a level above 20 us/cm, but that too is adequate.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.