HCBBS Forum (English)
Submit Chemical Projects / Find Solutions
Amplify Your Requirements on a Broader Chemical Platform *Engineering · Technology · Equipment · Solutions*
Submit Request

Determination of ambient temperature when specifying requirements for instrument bidding

2018-10-17View Original

Thread Content

Available meteorological data: Annual average temperature: 13.5℃; Annual highest temperature: 42.5℃; Annual lowest temperature: -19.0℃; Average winter temperature: -2.0℃; Average summer temperature: 26.9℃. Does the ambient temperature for the instruments refer to the range of -2.0~26.9℃?
Reply #22018-10-17
This is general data, and it is correct to provide it. But it is best to specify the requirements for the specific location separately; the data sheet also includes information on the medium temperature, which can affect the selection of the instrument as well
Reply #32018-10-17
For normal operating conditions, there is no need for such strict accuracy; some error is inevitable
Reply #42018-10-17
I understand that the ambient temperature for the instrument is between -19 and 42.5°C; it is necessary to take into account the fact that the instrument may malfunction when the ambient temperature is low
Reply #52018-10-18
I think the ambient temperature of the instrument is -19 to 42.5℃
Reply #62018-10-18
The ambient temperature for the instrument is between -25 and 60°C, and most devices can meet this requirement.
Reply #72018-10-18
Unless you go to Siberia, in normal areas the temperature is around -25 degrees; however, with more extreme weather conditions in recent years, temperatures as low as -40 degrees could be a possibility.
Reply #82018-10-18
For industrial field instruments, when there is no display screen, the design temperature of the instrument’s circuit board should generally be in the range of -40 to 85°C. When a display screen is present, the design temperature of the circuit board should generally be between -20 and 60°C... There are also display heads that can operate at low temperatures of -30°C, -40°C, or -55°C, but these are not common; those capable of operating at -55°C are also very expensive.
Reply #92018-10-18
It should be noted, however, that the accuracy of the instrument is determined by the nominal temperature of 20°C, which refers to the condition in which the instrument operates at an environment temperature of 20°C... Refer to relevant **standards and international standards. When the actual operating temperature of the instrument deviates from the nominal temperature, especially by a significant amount, it is necessary to take into account the temperature error of the instrument – often this temperature error is much greater than the instrument’s basic precision error
Reply #102018-10-19
:learned handshake*

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.