2015-07-27View Original

Thread Content

Capillary column, FID detector. H2: 30 mL/min; Air: 300 mL/min; Temperature: 250°C; Tail gas flow: 15 mL/min. The detector shuts down after operating for one or two days; the flow rates of all gas supplies are normal. Please help
Reply #22015-07-27
Reduce the tail gas flow rate, and increase the air and hydrogen flow rates
Reply #32015-07-27
See the link for reference: http://bbs.foodmate.net/thread-461109-1-1.html
Reply #42015-07-28
Increase the auxiliary air flow slightly, and also increase the hydrogen flow a bit; this may raise the signal baseline a little, but the flame will be more stable. Also check whether there is carbon buildup in the burner; if so, clean it according to the manual instructions.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.