Why is the accuracy poor at low flow rates for flow meters? This is considered a feature of the flow meter.
The measurement methods and approaches should be determined based on the requirements. Firstly, there is a low demand for measuring small flow rates, and even less demand exists for high precision in such cases. As a result, most flow meters do not take small flow rates into account, or they are simply not usable for such situations.
Meters all have a sensitivity requirement; for example, it should be at least 3%, as accuracy of flow measurement will be very poor if it is below 3%. What is relatively easy to understand are the pressure gauges on site; the scale around the zero point does not start at 0, but for example at 0.2.
When the flow rate is low, the resulting flow signal is weak; every type of flow meter has difficulties dealing with such weak signals, and detecting them is currently a bottleneck for flow meters.
Any medical device has a range of applications; a medicine that can cure all diseases is obviously a fake drug – isn’t that quite obvious? Using a meter stick to measure the dimensions of precision watch parts is certainly not appropriate; you need to use a micrometer instead. If it is anticipated that low flow rates will need to be measured frequently, then when selecting a flow meter, one should choose one that is suitable for the range of low flow rates that are frequently measured. Alternatively, a bypass can be installed in the pipeline: a flow meter with a smaller diameter suitable for low flow rates can be used there, while a flow meter designed for higher flow rates can be installed in the main pipeline. Thus, the appropriate flow meter can be chosen depending on the flow rate involved.
Thank you for your reply. For example, in the case of thermal flowmeters, we generally say that the accuracy is 1% of the reading plus 0.5% of the range. When using calculation software, the accuracy is lower at low flow rates. So how is this accuracy determined?
Didn’t you already say it’s calculated using software?
Accuracy – what exactly is accuracy? For most instruments, accuracy refers to the full-scale accuracy (FS). The formula for calculating this is: FS = Maximum error / Full scale of the instrument * 100%. The full scale of an instrument is determined by the manufacturer of that instrument; it is an attribute of the instrument itself, not related to the range of measurements you can take. When dealing with low flow rates, the formula for calculating the measurement error should be: Error = Maximum error / Range of measurement * 100% = FS * Full scale of the instrument / Range of measurement * 100%. Do you think this value is high?~~
It mainly depends on what type of flow is involved; each type of flow can be analyzed in detail. For example, with vortex flow meters, a low flow rate means a low velocity. When the velocity drops to a certain level, the Karman vortices become less distinct, which affects the detection accuracy; Another example is that differential pressure flow meters generate a very small pressure difference at low flow rates, which makes it difficult for the sensor to detect accurately. I’m not sure which type of flow meter you’re referring to. We can analyze the reasons by looking at its principle! If we don’t understand, let’s discuss it again