e+h ultrasonic L-flow meter reports error 91W, C281 (initialization)

2016-10-27View Original

Thread Content

The e+h ultrasonic flow meter model 91W is showing an error code C281 (initialization). The value keeps fluctuating – what should I do? Even after turning off the power, the problem persists; is there any solution? Please help
Reply #22016-11-02
Mu Weiqiang, I can run into you even here. Come down and have a drink
Reply #32018-05-02
How did you resolve this error code?
Reply #42018-05-03
First, grind the pipe again to remove rust. Second, apply some more coupling agent. Third, whether the pipeline parameters are entered correctly in the table header leads to inaccurate installation locations. Fourth, is the installation location causing bubbles to form in the pipes from time to time? All of the above factors can lead to C281, which means initialization; in practice, it involves finding a stable signal again.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.