An 1151 pressure transmitter has a range of 0–10 kPa abs. It is calibrated using a standard pressure gauge with a range of 0–100 kPa abs. If the local atmospheric pressure is 98 kPa abs, then when the transmitter outputs 12 mA, the corresponding input pressure value is 7 kPa abs. Please explain this. If you can’t figure out these relationships, give some guidance
Reply #22018-06-04
12 ma corresponds to 5 kPa abs; it’s possible that your gauge is inaccurate
Reply #32018-06-05
Xiaoxinxin, what does ABS mean? What do you need to do? For a differential pressure gauge, the low-pressure side should be connected to the ambient atmosphere; there’s no need to worry about the value of the ambient atmospheric pressure. Both the high-pressure side of the differential pressure gauge and the standard gauge should be connected to the input pressure, and then the gauge can be calibrated using gauge pressure. The range of the standard gauge you use should be much larger than that of the differential pressure gauge – this isn’t appropriate. You need to use a standard gauge with a smaller range, such as 10 kPa or 30 kPa. Otherwise, the accuracy of the calibration will be very poor
Reply #42018-06-05
Enter 0 kPa for a calibration of 4 mA, and enter 10 kPa for a calibration of 20 mA. After adjusting these two values accurately, it may be necessary to make several adjustments before achieving the correct setting. Once that is done, observe the current output at characteristic pressures such as 2.5 kPa, 5 kPa, and 7.5 kPa
Reply #52018-06-08
Just because it can’t be done doesn’t mean the standard table is inaccurate. Since it is about calibrating the transmitter, the standard gauge should be used as a reference (once this standard gauge is chosen, there is no need to consider whether it is accurate or not). It could also be due to an error in selecting the standard table; if the standard table shows 7 kPa, it means that a pressure signal of 7 kPa is being sent to the transmitter, and if the transmitter outputs 12 mA, then it indicates that the transmitter is inaccurate and needs to be adjusted.
Reply #62019-01-30
This shows the relationship between absolute pressure and gauge pressure; absolute pressure is equal to atmospheric pressure plus gauge pressure. Since the local atmospheric pressure is below 100 kilopascals, a compensation of 2 kilopascals is needed. 5 + 2 = 7, which corresponds to 12 milliamps

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.