Differential pressure gauge replacement

2018-07-13View Original

Thread Content

The diaphragm of a differential pressure gauge of type EJA110A on site is damaged (with a range of 0-5 kPa), and no suitable new gauge can be found. Is it possible to replace it with an EJA110E gauge (with a range of 0-100 kPa)?
Reply #22018-07-13
If it can be adjusted to 0–5 kPa, it can then be replaced.
Reply #32018-07-13
I was wondering if it’s possible to use 475 to directly change the range on the gauge to 0-5 kPa First replacement
Reply #42018-07-13
Try it and you'll see....
Reply #52018-07-13
It can be changed. . . If you want uncertainty, go ahead and mess things up however you like. . . Keep going, keep going. :lol. .
Reply #62018-07-13
There is no model EJA110E; it should be the EJA110A-E -M, with a range of 0-100 kPa. The one that is broken is the EJA110A-E-L, with a range of 0-10 kPa. The minimum operating range for the EJA110A-E -M is 0-1 kPa, but the client needs a range of 0-5 kPa, which is still feasible – the range can be adjusted using the manual controller.
Reply #72018-07-15
It depends on his 20% ratio; if it’s 100 to 1, then it’s fine, but if it’s 10 to 1, then nothing can be changed
Reply #82018-07-15
It’s possible to adjust the range; it’s best to have it calibrated at the metrology laboratory
Reply #92018-07-15
The pressure difference is relatively small; after installation, it is necessary to perform a zero-point calibration with air
Reply #102018-07-16
Well, it can definitely be changed. Checked the parameters: faulty gauge EJA110A-A-L, range 0-10 kPa, accuracy 0.075% ; Replacement table EJA110A-E -M, 0-100kpa, accuracy 0.055%. When measuring 0-5 kPa, change the range to 0-10 kPa; the accuracy remains at 0.055%.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.