The unit’s storage tank is equipped with a double-flange level gauge; the lower flange is clogged like this – can it still display the liquid level? As shown in the image
Reply #22019-10-10
It is recommended to use plug-in double flanges; the insertion depth should be determined based on the flanges of your tank
Reply #32019-10-10
If conditions permit, it is recommended to consider using a radar level gauge
Reply #42019-10-10
It’s blocked, so the displayed liquid level is inaccurate.
Reply #52019-10-10
Pressure transmission will definitely be affected; it is not allowed
Reply #62019-10-10
For relatively static equipment, if there is sufficient funding, replacing it with a radar level gauge would cost around over 10,000 yuan for the anti-corrosion measures.
Reply #72019-10-10
A float level gauge can also be used; there are anti-corrosion versions of them, but no solid materials should be present inside, as this could cause blockages.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.