Ultrasonic level gauge E+H FMU40

2020-02-28View Original

Thread Content

Please, experts! ! Fix it! Thank you! ! I have an E+H ultrasonic radar level gauge. After it was installed, the 475 signal couldn’t be connected to the gauge’s display; it was only possible to carry out the calibration after connecting it to the DCS cabinet. Now that the tank is already full, the gauge still shows 0.0%. How can this problem be solved? I hope someone experienced in this area can help me out. Thank you! !
Reply #22020-02-28
A milliampere circuit is connected in series with a 250-ohm resistor; can’t communicate either when 475 is connected across the resistor?
Reply #32020-02-28
When connected to the field device, there is likely no stray resistance; when connected to the DCS, the safety barrier introduces resistance. If the meter is functioning properly, a reading of 0% may indicate an empty value.
Reply #42020-02-28
Check the instructions to set the empty height, full height, and blind area
Reply #52020-02-28
Yes, it wasn’t set during debugging

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.