Reaction lag of pure oxygen analyzer

2016-02-24View Original

Thread Content

For oxygen products in air separation, the ABB EL3020 Magnos206 pure oxygen analyzer shows a concentration range of 98.0~100.0%; there is severe reaction lag. Has anyone encountered this issue? Calibration with standard gas is all normal. Is it a installation issue? The volume is fine as well, and the float ball is working properly.
Reply #22016-02-24
What method is used for this analysis? In ranges with such high oxygen concentrations, it is not recommended to use electrochemical or magnetic oxygen methods; it is better to employ chromatography to analyze the impurity content and calculate purity using the difference method.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.