How to calibrate a weighbridge after replacing all of its weighing sensors?
Reply #22016-05-26
First, roll back and forth on the scale several times with a load of around 80% of its capacity, to allow the sensor to make repeated contact with the foundation. Next, angle deviation adjustment is performed by placing a weight with a capacity of 1/10 on top of the sensor; the weight should be placed as centrally as possible to bring the angle deviation within the acceptable error range. After the angle deviation has been adjusted, a weight with the maximum capacity is used to carry out range adjustment. After one such cycle, the adjustment is complete.....

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.