How to deal with fluctuations in the flow rate of Emerson mass flow meters between -990 kg/h and 1000 kg/h during normal operation?
Reply #22017-02-05
Let’s try using 475 first to see if there are also fluctuations
Reply #32017-02-06
There are two aspects to this: one is the sensor, and the other is the core processor. Use 475 to conduct a diagnosis; if no issue can be detected, try replacing the core processor to see if that helps
Reply #42017-02-06
Check for interference, pipe vibration.
Reply #52017-02-06
Your question is unclear; is it at midnight? Still during measurement?
Reply #62017-02-06
The error code is A102; it has been addressed, but the issue still persists
Reply #72017-02-06
The most likely possibility is two-phase flow. If it functions normally when completely emptied, then it must be a two-phase flow. Also, how did you rule out a102?

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.