Thermal resistor is submerged in water; the thermometer gives inaccurate readings. Help!

2015-12-08View Original

Thread Content

Water entered the sheath of the thermistor thermometer, causing it to display 0°C; normal readings were obtained after replacing the thermistor. The thermistor is of the PT1000 type; after water entered it, its resistance was measured at 500 ohms. The next day, this resistance increased gradually to 900 ohms. Even after it was reinstalled in the thermometer, it still indicated 0°C. What could be the reason for this? Can this thermistor still be repaired for further use? Thank you!
Reply #22015-12-08
It’s new and not expensive; just buy a new one. . . . . . . . . . . .
Reply #32015-12-08
How much is one 2-meter long PT1000 (Grade B) wire? What I mainly want to ask is the reason and whether it can be fixed? Thank you
Reply #42015-12-08
Thermoresistors that fail the calibration should be replaced.
Reply #52015-12-08
When water flows in, it’s as if the resistance is short-circuited; a resistance value of zero is correct.
Reply #62015-12-08
It’s simple: contact your supplier. He can repair it for you, but generally, when a thermal resistor is damaged, it’s replaced rather than repaired.
Reply #72015-12-08
It costs a few hundred yuan; we can work together if there’s a need
Reply #82015-12-08
This post was last edited by dennyshen2549 on 2015-12-9 at 17:36. You can check the division value of PT1000 and compare it; at 0 degrees, it is around 1000 ohms
Reply #92015-12-08
Measure after drying the thermistor; if that doesn’t work, replace it with another one.

Submit a Project

**Looking for Chemical Technology, Equipment & Solutions?** No Registration Required Broader Platform Exposure | Global Chemical Service Provider Connections

Submit Request — Free Consultation

Disclaimer

This is an automated machine translation of the original thread. Some technical terms may have inaccuracies; the original text shall prevail. Click "View Original" at the top right to access the source page, which supports IP-based automatic real-time language translation. Please watch out for contact details and sales inducements to prevent fraud. All content and translations are for reference only, representing solely the poster's personal views. For enquiries, email service@hcbbs.com.